Old 2nd Jun 2022, 11:56 am   #181
Alan Bain
Pentode
 
Join Date: Jan 2012
Location: Heswall, Merseyside, UK
Posts: 165
Default Re: Spectrum Analyser TR4172

By some micacle I saw the capacitor which had pinged out of the black blob on the board. It is connected between base and ground of the "bias" transistor. I was able to solder leads on and measure on a bridge 10nF in a 0805 package. Leaving this here in case of use to anyone in future.
Alan Bain is offline  
Old 3rd Jun 2022, 2:32 pm   #182
Matt kd4pbs
Tetrode
 
Join Date: Mar 2021
Location: Raleigh, North Carolina, USA.
Posts: 62
Default Re: Spectrum Analyser TR4172

Good find! For sure, far less than I thought it would be.
I do hope you get this going, Alan. It's a wonderful SpecAn... when it works. Super low noise, great filtering, and a high resolution, even by today's standards. What's not to love? Oh yeah... the age-related issues... I forgot
Matt kd4pbs is offline  
Old 22nd Jun 2022, 9:17 am   #183
Alan Bain
Pentode
 
Join Date: Jan 2012
Location: Heswall, Merseyside, UK
Posts: 165
Default Re: Spectrum Analyser TR4172

Yes, replacing smelly electrolytics in the RF section is a fairly slow process. Almost back able to do some actual RF testing!
Alan Bain is offline  
Old 25th Jun 2022, 9:05 pm   #184
Alan Bain
Pentode
 
Join Date: Jan 2012
Location: Heswall, Merseyside, UK
Posts: 165
Default Re: Spectrum Analyser TR4172

Back together and starting to test to find what I forgot to connect - running into a few manual discrepancies; manual makes reference of a T.G CNTR button but when I look where that should be I find TUNED AMP. Is this just a new name for the same thing (as in count the actual signal not just calculated centre + offset)?
Alan Bain is offline  
Old 26th Jun 2022, 8:16 pm   #185
Alan Bain
Pentode
 
Join Date: Jan 2012
Location: Heswall, Merseyside, UK
Posts: 165
Default Re: Spectrum Analyser TR4172

Here is the repaired first IF bias board (after and before pictures!). I have to admit that it was tricky. The board material conducts heat away very effectively and the gold layer appears to be pure gold (so no nickel in the alloy) and doesn't wet with solder. I found I had to use a powerful iron to be able to work quickly. The resistive layer underneath does wet with difficulty. I was very reluctant to use anything corrosive to clean the surfaces given the comments in this thread about glue cure by products damaging trace adhesion. New parts deliberately not straight to avoid the damaged trace sections. One trace to decoupling 10n cap replaced by a wire (single strand of 7/0.2 hookup wire).

Afterwards now drawing 13mA instead of 22mA+ and voltages look sane. Time for more analyser testing.
Attached Thumbnails
Click image for larger version

Name:	DSC_5385_bal.jpg
Views:	107
Size:	88.4 KB
ID:	259873   Click image for larger version

Name:	DSC_5381_before.jpg
Views:	99
Size:	64.7 KB
ID:	259874  
Alan Bain is offline  
Old 27th Jun 2022, 3:43 pm   #186
Alan Bain
Pentode
 
Join Date: Jan 2012
Location: Heswall, Merseyside, UK
Posts: 165
Default Re: Spectrum Analyser TR4172

After testing further I've noticed that the directional coupler is terminated in 50ohm on input side but measures about 600ohms looking back from C1. While I don't have any values any design for such a coupler that I have seen has had both ports terminated equally. Fortunately I have some precision 0402 50ohm terminating resistors, but does anyone have any ideas of a flux that allows the gold to be wetted?
Alan Bain is offline  
Old 15th Jul 2022, 4:30 pm   #187
Matt kd4pbs
Tetrode
 
Join Date: Mar 2021
Location: Raleigh, North Carolina, USA.
Posts: 62
Default Re: Spectrum Analyser TR4172

Can't help you there, Alan. I noticed the same thing with the insanely difficult to solder properties of the substrate, and with my iron at a heat setting that would overcome the thermal conduction from the board, it would melt the traces.
I think I do remember having the same issue that you see... the directional coupler/filter not being equally terminated.
Ultimately, after bodging the SA enough that I got it working and calibrated, after a few weeks I noticed it went deaf again. Likely a leg of the mixer corroded again. I have since pulled it from the shelf and set it in the morgue awaiting either the day it gets carted to the dump or I find the required parts to make it right.
Matt kd4pbs is offline  
Old 16th Jul 2022, 8:09 pm   #188
Alan Bain
Pentode
 
Join Date: Jan 2012
Location: Heswall, Merseyside, UK
Posts: 165
Default Re: Spectrum Analyser TR4172

So having mended the IF amp, remaining deafness hard to trace without an SA which covers 2-3GHz. Thanks to a kind loan, the deafness traced to one of the stripline 2.046GHz bandpass filter sections which is an approximately 60db attenuator. Careful inspection shows one section of stripline has no gold so no great surprise and I should have noticed sooner! For the moment I have put in a piece of 50ohm semi-rigid coax to bypass this section and with this bodge I can get it to calibrate.

I vaguely recollect doing the J-admittance inverter calculations to design one of these filters for a Chebychev response and I have a quick response measurement from the other so that shouldn't be so hard although this looks like an Alumina substrate around 0.72mm thick ... don't think the cheap Chinese PCB people do Alumina, or Rogers 3010! And I also seem to remember some iterative length adjustment was needed to the lines....

Click image for larger version

Name:	FilterResponse.png
Views:	57
Size:	8.5 KB
ID:	261070
Alan Bain is offline  
Old 23rd Sep 2022, 7:39 pm   #189
Alan Bain
Pentode
 
Join Date: Jan 2012
Location: Heswall, Merseyside, UK
Posts: 165
Default Re: Spectrum Analyser TR4172

I thought an update on what I found with my TR might be of interest as my repairs have been very slow....

1. Hairline crack in the trace in the attenuator (just after the AC/DC coupling selection). This acted as a capacitor at high frequency, but caused much extra attenuation at low and I didn't actually find it until I put the unit on the VNA which produced a very capacitative S21. DC measurement confirmed the problem. Soldered a piece of wire over the break under the microscope.

2. PLL fault for 101/100MHz PLL caused by two problems. Someone had "undone the screws" in the xtal harmonic multiplier (alignment using another SA cured this) and there was a leaky NP electrolytic in the loop filter. Unfortunately the manual and the boards I have don't match - for example my VCO modules are 3 pin devices with the tuning and power combined, but the KO4BB manual shows 4 pin devices also mine uses a MC4044 PFD whereas manual shows JK flipflops and an AND gate. As a result the alignment instructions don't work (it is necessary to know what voltage to apply with the loop opened before tuning). Mostly working but would like to know the correct alignment procedure. There seem more operating manuals than service. If I were to guess from the sequence of parts numbers I'd say my board is earlier.

3. 2.046GHz filter that doesn't work in the IF interface. As suspected the gold traces had lifted leaving the resistive material below (must have happened long before it came into my hands). I replaced by a piece of fibreglass board with a 50 ohm microstrip for now and have redesigned a 4 pole bandpass filter onto RO3210 which needs fabricating (with proper photo-etching not dubious toner transfer).

4. Bias network for 1st IF amp not working. Added external SMD resistors and replaced bias transistor until correct quiescent current.

With these done it self-calibrates and works fine (although I am sure IM performance is impaired by the "special" filter). It's a very pleasant SA to use and feels fairly modern. I haven't tried the impedance mode yet.
Attached Thumbnails
Click image for larger version

Name:	DSC_5441.jpg
Views:	83
Size:	95.2 KB
ID:	265170  
Alan Bain is offline  
Old 23rd Sep 2022, 10:45 pm   #190
G0HZU_JMR
Dekatron
 
Join Date: Sep 2010
Location: Cheltenham, Gloucestershire, UK.
Posts: 3,077
Default Re: Spectrum Analyser TR4172

Good work! My TR4172 also suffered a microstrip failure in the attenuator section.

If you do experiment with the impedance mode, turn down the CRT brightness first. Otherwise, it can draw a pin sharp stationary dot on the smith chart that can be frighteningly bright. However, I've not burnt the CRT on mine so maybe it is fairly immune to burn. The calibration process for the impedance mode is quite clunky and time consuming but on the few occasions I tried to use it, the performance was much better than I was expecting. However, I think a modern nanovna is going to be easier to use and will usually give much better performance for impedance measurements.

I spent many years using this TR4172 at work, so I grew quite fond of it. When the chance came to save it from disposal, I bought it from the company for the cost of a small charitable donation. The strongest points of the TR4172 were the second harmonic intercept and the LO phase noise at 100kHz offset. The mixer IP3 was impressive too at about 21dBm across HF and into VHF.

I mainly used it for VHF and UHF synthesiser design back in the early 1990s and also for distortion measurements. The KE5FX plotter software can work with it via GPIB but it is very fiddly to set up on a TR4172 that has the options that mine has. It's not an issue with the KE5FX SW, I think it depends if the TR4172 has the enhanced GPIB capability fitted. No-one at work could ever get this TR4172 to plot with an old school HP plotter but it isn't faulty as such.
__________________
Regards, Jeremy G0HZU
G0HZU_JMR is offline  
Old 25th Sep 2022, 9:10 pm   #191
Alan Bain
Pentode
 
Join Date: Jan 2012
Location: Heswall, Merseyside, UK
Posts: 165
Default Re: Spectrum Analyser TR4172

It seems the CRTs do burn; mine has a very pronounced image of the rectangular graticule even when switched off!
Alan Bain is offline  
Old 10th Feb 2023, 11:15 pm   #192
Alan Bain
Pentode
 
Join Date: Jan 2012
Location: Heswall, Merseyside, UK
Posts: 165
Default Re: Spectrum Analyser TR4172

Leaving this here in case someone else finds it useful. I measured one of the 2.046GHz filter boards in the undocumented 1st IF amp using a toolmaker's microscope. I was measuring the underlying resistive layer where the gold had vanished.

The filter is a clearly a parallel coupled transmission line type of 3rd order. Track form is -B-A-A-B- with 50 ohm input and output microstrip on Alumina substrate (25.6 thou substrate, gold thickness 0.39 thou). Assume Dk of 9.8 for Alumina (I don't have any way to measure - suggestions welcome) to calculate the odd and even mode characteristic impedances from the dimensions:

B section strip width 21.0 thou, spacing 11.4 thou so Z_{oo}=39.31 ohm and Z_{oe}=66.8 ohm

A section strip width 23.9thou, spacing 46.5 though so Z_{oo}=47.19 ohm, Z_{oe}=54.33 ohm

So whole filter can be reduced to 4 admittance "J" inverters with half-wave lines in between (which are seem as a parallel resonant circuit, susceptance sensitivity is pi Y_0/2).

---J_{0,1}--lambda/2----J_{1,2}----lambda/2----J_{2,3}---lambda/2---J_{3,4}---

From the odd and even impedances the J (or K inverter params) are quickly calculated as
J_{0,1}/Y_0 = J_{3,4}/Y_0 = 0.28
J_{1,2}/Y_0 = J_{2,3}/Y_0 = 0.071

where Y_0 = 1/Z_0 (characteristic admittance). With rather more algebraic pain, by guessing a Chebychev low pass prototype transformed to a bandpass filter. The Normalised Low pass component values which would need the above J inverters is g_1=0.923, g_2=1.126, g_3=0.923 (for a normalised frequency of 1 rad/s and g_0=g_4=1 for the terminations).

These are the values computed for a 0.06db ripple 3rd order Chebychev and the scaling then corresponds to a bandwidth of adjusted to 94MHz (assuming center frequency 2.046GHz from the length of lines).

So now it remains to lay lines out on some RO3210 as alumina seems unobtainable for the same Z_{oo}, Z_{oe}. Thankfully no through holes to worry about, only making dry photoresist stick to the copper!
Alan Bain is offline  
Old 27th May 2023, 1:49 pm   #193
Alan Bain
Pentode
 
Join Date: Jan 2012
Location: Heswall, Merseyside, UK
Posts: 165
Default Re: Spectrum Analyser TR4172

Attached are Gerber files for the first of the two 2.046GHz IF band pass filters in the 1st IF module oif the 4172. This is redesigned to be used on 25 thou Rogers RO3210 substrate (epsilon_R=10.8) rather than alumina. Tested in Sonnet but awaiting fabrication which sadly doesn't seem possible via the usual low cost places on this substrate. At least there are no vias so DIY methods should work!

The second filter is a virtually identical layout apart from the location of the input and output pads.
Attached Files
File Type: zip Gerber-2046.zip (4.7 KB, 15 views)
Alan Bain is offline  
Closed Thread

Thread Tools



All times are GMT +1. The time now is 10:20 am.


All information and advice on this forum is subject to the WARNING AND DISCLAIMER located at https://www.vintage-radio.net/rules.html.
Failure to heed this warning may result in death or serious injury to yourself and/or others.


Powered by vBulletin®
Copyright ©2000 - 2024, vBulletin Solutions, Inc.
Copyright ©2002 - 2023, Paul Stenning.